Cambridge International A Level Physics 9702
Core idea: medical images are reconstructed from measured signals—ultrasound echoes, transmitted X-rays or coincident gamma photons—not direct photographs of internal structure.
Exact syllabus boundary
24.1 ultrasound generation, echoes, acoustic impedance, reflection and attenuation; 24.2 X-ray production, imaging, attenuation and CT; 24.3 tracers, positron annihilation and PET coincidence. MRI and therapy are outside Topic 24.
Piezoelectric transducer
Medical imaging extracts information from how waves or radiation interact with tissue.
Ultrasound uses reflected sound; X-ray imaging uses transmission and attenuation; PET uses photons from positron–electron annihilation.
Role
Piezoelectric action
transmit
alternating p.d. makes the crystal change shape/vibrate, producing ultrasound
receive
returning pressure variations deform the crystal and generate an alternating e.m.f.
Wording trap
Do not reverse the two effects: applied p.d. causes deformation; deformation causes e.m.f.
2
Pulse echoes and depth
A short pulse is followed by a listening interval. At each boundary, part of the pulse may reflect and return to the transducer.
$d=c\Delta t/2$
The factor 2 is essential because the pulse travels to the boundary and back. For two boundary echoes, layer thickness is the difference between the two calculated depths—or $c(t_2-t_1)/2$ when the speed in that layer is known.
Worked echo
For $c=1540\,\mathrm{m\,s^{-1}}$ and round-trip time $52\,\mu\mathrm s$, $d=1540(52\times10^{-6})/2=4.00\,\mathrm{cm}$.
Common mistake
A later echo identifies a greater travel time, not automatically a stronger boundary. Echo amplitude also depends on reflection and attenuation.
3
Acoustic impedance and reflection
For ultrasound echoes, the recorded time is usually a round trip.
Reflection strength depends on acoustic-impedance mismatch.
In X-ray images, different attenuation produces contrast; CT combines measurements from many directions to reconstruct sections.
$Z=\rho c$
Acoustic impedance depends on both density and ultrasound speed. At normal incidence, the reflected intensity fraction is
Under an ideal boundary assumption with no absorption at the boundary, transmitted fraction $T=1-R$.
Coupling medium
A coupling layer reduces the large impedance mismatch that an air gap would produce, allowing more ultrasound to enter the next medium.
Percentage trap
The formula already gives an intensity fraction. Multiply by 100 only when a percentage is requested.
4
Ultrasound attenuation
$I=I_0e^{-\mu x}$
Attenuation is the reduction of wave intensity as energy is absorbed or scattered while the ultrasound travels through a medium.
$\ln(I/I_0)=-\mu x$, $x_{1/2}=\ln2/\mu$
Situation
Method
one uniform layer
use its $\mu x$ in the exponent
several layers
multiply attenuation factors, equivalent to adding $\mu_i x_i$
echo path
include outward and return travel distances
Unit trap
If $\mu$ is in cm⁻¹, use $x$ in cm. Reflection at a boundary and attenuation through a medium are separate losses.
5
X-ray production and minimum wavelength
A heated cathode releases electrons.
A large p.d. accelerates them through a vacuum.
They strike a metal target and decelerate rapidly, producing X-rays.
The shortest wavelength is the maximum photon energy case.
$eV=E_{k,max}=hf_{max}=hc/\lambda_{min}$
Increasing accelerating p.d. increases maximum photon energy and decreases $\lambda_{min}$. Increasing tube current increases electron rate and X-ray intensity, but does not change $\lambda_{min}$ when V is unchanged.
Energy trap
$eV$ is the maximum energy available from one electron; it does not mean every emitted X-ray photon has that energy.
6
X-ray attenuation, contrast and CT
$I=I_0e^{-\mu x}$
Different materials attenuate X-rays by different amounts, so the detector receives different intensities and the image has contrast.
CT sequence
Record X-ray attenuation projections from many angles around one cross-section.
A computer combines the projections to reconstruct one 2D slice.
Repeat at neighbouring positions along the body/object.
Combine the slice stack to form a 3D representation.
Sequence trap
One projection is not a CT slice. A slice is reconstructed from many angular projections.
7
PET tracers and coincidence detection
A tracer contains a positron-emitting isotope and participates in the process being studied, so its distribution can be detected.
The nucleus emits a positron.
The positron loses energy and annihilates with an electron.
Approximately opposite gamma photons are produced to conserve momentum.
Opposite detectors record nearly simultaneous arrivals.
Each detector pair identifies a line of response; many events reconstruct tracer concentration.
$E_\gamma=m_ec^2=0.511\,\mathrm{MeV}$ for each photon when initial kinetic energy is negligible.
Localisation
A detector pair alone gives a line, not a unique point. Arrival-time differences can refine position along that line; many coincidences are combined for the image.
8
Paper 4 technique & master equation sheet
Representation-to-method map
Echo times → mark round trip → divide by 2.
Boundary data → calculate both impedances before the reflection coefficient.
Layer data → make $\mu x$ dimensionless and separate attenuation from reflection.
X-ray tube → one electron gains $eV$; maximum photon has $hc/\lambda_{min}$.
CT diagram → projections, then slice, then volume.
PET event → energy, opposite momentum, coincidence line, many-event reconstruction.
Master equations
echo depth
$d=c\Delta t/2$
acoustic impedance
$Z=\rho c$
reflection coefficient
$R=[(Z_2-Z_1)/(Z_2+Z_1)]^2$
attenuation
$I=I_0e^{-\mu x}$
X-ray limit
$eV=hc/\lambda_{min}$
annihilation photon
$E_\gamma=m_ec^2$
photon momentum
$p=E/c$
Final traps
divide echo distance by 2
$R$ is a fraction before ×100
match units in $\mu x$
tube current does not set $\lambda_{min}$
many CT projections make one slice
one PET pair defines a line
Scope
These notes explain the examination physics of imaging systems; they are not clinical guidance.